Everything below concerns Chromatogram. We keep the language plain, cite what the science says, and separate well-supported claims from open questions.
Updated 2026-04-25. Numbers and descriptions here follow the published literature rather than marketing material.
Quality control laboratories use HPLC to check identity, purity, concentration, and stability of raw materials and finished products. A validated method specifies the column, mobile phase, flow rate, detection wavelength, injection volume, and run time. Samples are prepared and compared against reference standards of known concentration. The resulting chromatogram provides quantitative data, such as assay values and impurity levels. This approach is common in pharmaceutical, food, environmental, and industrial testing where consistent measurements are required.
Method validation demonstrates that an analytical procedure is suitable for its intended purpose. Typical validation characteristics include accuracy, precision, specificity, linearity, range, limit of detection, limit of quantitation, and robustness. Regulatory guidance from bodies such as the International Council for Harmonisation and the United States Pharmacopeia outlines expectations, though specific criteria depend on the product and method. System suitability tests are run before sample analysis to confirm resolution, peak symmetry, column efficiency, and injection repeatability. Failure of these checks can invalidate a batch of measurements.
Separation modes differ by the chemistry of the stationary phase and the composition of the mobile phase. Reversed-phase testing uses a nonpolar column and polar solvents, making it common for pharmaceutical, environmental, and food analytes. Normal-phase testing uses a polar column and nonpolar solvents for compounds that are poorly retained in reversed-phase systems. Ion-exchange and ion-pair methods separate charged species, while size-exclusion methods sort molecules by hydrodynamic volume. Gradient elution changes solvent strength over time to resolve complex mixtures, and isocratic elution holds solvent composition constant for simpler assays.
Key performance measures include retention time, peak area, peak height, resolution, tailing factor, and plate count. Retention time helps identify a peak under fixed conditions, but confirmation often requires a second method or detector. Peak area and height relate to concentration through calibration curves, which may be linear or nonlinear depending on the detector response. Resolution describes separation between adjacent peaks, while tailing factor and plate count describe peak shape and column efficiency. Performance checks verify these values before and during a run to confirm that the instrument is performing within limits.
High-performance liquid chromatography testing separates components of a liquid sample by forcing a mobile phase through a packed column. The stationary phase inside the column interacts with analytes to different degrees, so each compound exits at a characteristic retention time. A pump delivers solvent at controlled flow and pressure, while an injector introduces a precise sample volume. Detectors such as ultraviolet-visible, fluorescence, refractive index, or mass spectrometric instruments record the separated bands. The resulting chromatogram provides qualitative and quantitative information about the mixture.
| Parameter | Typical acceptance criterion | Notes |
|---|---|---|
| Resolution | ≥ 1.5 | Baseline separation of adjacent peaks |
| Tailing factor | ≤ 2.0 | Peak symmetry measure |
| Theoretical plates | > 2000 | Column efficiency indicator |
| Injection repeatability | ≤ 2% RSD | Relative standard deviation for replicate injections |
| Linearity | r² ≥ 0.995 | Calibration curve over the working range |
Detection commonly uses ultraviolet-visible absorbance, fluorescence, refractive index, or mass spectrometry. Ultraviolet detection depends on molecular chromophores that absorb light at specific wavelengths. Mass spectrometry provides mass information and sensitive quantification, often after electrospray ionization. Before sample batches, performance checks examine resolution, elution time repeatability, peak symmetry, and plate count. Matrix effects and co-elution remain recognized uncertainties; formal validation studies and orthogonal detection help address them. Detector choice depends on analyte properties and required sensitivity.
High-performance liquid chromatography, or HPLC, separates dissolved compounds by passing a liquid mobile phase through a packed column. Components distribute differently between the stationary phase and the moving liquid, so they travel at different speeds and exit at different times. A detector records these eluting bands as peaks, and peak area or height relates to amount. The technique supports testing in pharmaceuticals, foods, environmental samples, and industrial chemicals. Quantification usually depends on calibration with known standards.
Several separation modes exist, including reversed-phase, normal-phase, ion-exchange, size-exclusion, and hydrophilic interaction liquid chromatography. Reversed-phase uses a nonpolar stationary phase with a polar mobile phase and is widely applied to small organic molecules. Gradient elution changes mobile phase composition during the run, while isocratic elution keeps it constant. Column chemistry, particle size, temperature, flow rate, and mobile phase pH all influence retention and resolution. Method development selects conditions that separate analytes from matrix components and from each other.
Instrumentation includes a solvent delivery system, an autosampler, a column oven, and one or more detectors. Reversed-phase columns with chemically modified silica are widely used, but normal-phase, ion-exchange, size-exclusion, and affinity modes exist for specific separations. Detectors may rely on ultraviolet absorbance, fluorescence, refractive index, or mass spectrometry. Column temperature, mobile phase composition, and flow rate are adjusted to improve resolution. System pressure is monitored because rising pressure can indicate column blockage or deteriorating packing.
Separation performance depends on particle size, pore size, column length, and the chemistry of the stationary phase. Smaller particles generally improve efficiency but require higher pressure and suitable instrumentation. The mobile phase often contains buffers and organic solvents that influence retention and selectivity. Testing labs select conditions based on the analytes, sample matrix, and required sensitivity. Method development frequently involves screening several columns and solvent mixtures before a final set of conditions is chosen.
HPLC testing is an analytical technique used to separate, identify, and quantify components in a liquid sample. It relies on a pressurized mobile phase that carries the sample through a column packed with stationary phase. Different compounds travel at different rates because of interactions with the stationary and mobile phases. The resulting signal versus time is a chromatogram. Peak position indicates identity under specified conditions, while peak area or height relates to amount.
Laboratories apply HPLC testing across pharmaceutical, food, environmental, and industrial chemistry. The method can measure active ingredients, impurities, additives, preservatives, and degradation products. Sample preparation often includes dilution, filtration, and sometimes extraction or derivatization. The choice of column, mobile phase, pH, temperature, and detector depends on the analytes and matrix. Results are compared with reference standards to assign identity and concentration. Method suitability is judged by resolution, precision, and accuracy.
Most routine HPLC testing uses reversed-phase columns, where the stationary phase is nonpolar and the mobile phase is a polar mixture such as water with an organic solvent. Analytes partition between the two phases according to polarity, size, and charge. Gradients that change solvent composition over time can separate compounds with broad retention ranges. Isocratic conditions keep solvent composition constant and suit simpler mixtures. The choice of column chemistry, pH, and temperature affects selectivity and peak shape.
Detection in HPLC testing commonly relies on ultraviolet-visible absorbance, fluorescence, refractive index, or mass spectrometry. UV detection is widely used because many organic compounds absorb light, but it requires a chromophore. Mass spectrometry provides mass-based identification and high sensitivity for trace analytes. Each detector has trade-offs in selectivity, cost, and compatibility with mobile phases. Quantification typically uses calibration curves prepared from reference standards. Results are reported as concentration, purity, or presence above a limit.
HPLC testing separates dissolved compounds by passing a liquid sample through a column packed with stationary phase. A pump delivers mobile phase at controlled flow, and the sample components interact differently with stationary and mobile phases. Compounds that spend more time in mobile phase elute earlier; those retained by stationary phase elute later. Detectors record elution as peaks, and peak area or height relates to amount. This mechanism underpins quantitative analysis of mixtures.
The urine test may be a chromatographic immunoassay or any of several other test formats, home-, physician's office-, or laboratory-based. Published detection thresholds range from 20 to 100 mIU/mL, depending on the brand of test. Early in pregnancy, more accurate results may be obtained by using the first urine of the morning (when urine is most concentrated). When the urine is dilute (specific gravity less than 1.015), the hCG concentration may not be representative of the blood concentration, and the test may be falsely negative. The serum test, using 2-4 mL of venous blood, is typically a chemiluminescent or fluorimetric immunoassay that can detect βhCG levels as low as 5 mIU/mL and allows quantification of the βhCG concentration. The hCG levels grow exponentially after conception and implantation. hCG levels typically peak around weeks 8-11 of pregnancy and are generally higher in the first trimester compared to the second trimester. The following is a list of serum hCG levels: LMP is the last menstrual period dated from the first day of the last menstrual period
Pseudoacromegaly is a condition with the usual acromegaloid features but without an increase in growth hormone and IGF-1. It is frequently associated with insulin resistance. Cases have been reported due to minoxidil at an unusually high dose. It can also be caused by a selective post-receptor defect of insulin signalling, leading to the impairment of metabolic, but preservation of mitogenic, signaling. Treatment options include surgery to remove the tumor, medications, and radiation therapy. Surgery is usually the preferred treatment; the smaller the tumor, the more likely surgery will be curative. If surgery is contraindicated or not curative, somatostatin analogues or GH receptor antagonists may be used. Radiation therapy may be used if neither surgery nor medications are completely effective. Without treatment, life expectancy is reduced by 10 years; with treatment, life expectancy is not reduced.
Announced on 8 May 2024, AlphaFold 3 was co-developed by Google DeepMind and Isomorphic Labs, both subsidiaries of Alphabet. AlphaFold 3 is not limited to proteins, as it can also predict the structures of protein complexes with DNA, RNA, post-translational modifications and selected ligands and ions. AlphaFold 3 introduces the "Pairformer," a deep learning architecture inspired by the transformer, which is considered similar to, but simpler than, the Evoformer used in AlphaFold 2. The Pairformer module's initial predictions are refined by a diffusion model. This model begins with a cloud of atoms and iteratively refines their positions, guided by the Pairformer's output, to generate a 3D representation of the molecular structure. The AlphaFold server was created to provide free access to AlphaFold 3 for non-commercial research. As of November 2025, the AlphaFold 3 research paper has been directly cited more than 9,000 times.
The MCM6 gene, MCM6, is expressed at very high level. MCM6 contains 18 introns. There are 2 non overlapping alternative last exons. The transcripts appear to differ by truncation of the 3' end, presence or absence of 2 cassette exons, common exons with different boundaries. MCM6 produces, by alternative splicing, 3 different transcripts, all with introns, putatively encoding 3 different protein isoforms. MCM6 contains two of the regulatory regions for LCT, the gene encoding the protein lactase, located in two of the MCM6 introns, approximately 14 kb and 22 kb upstream of LCT. A substitution of thymine for cytosine in the first region (at -13910), in particular, has been shown to function in vitro as an enhancer element capable of differentially activating transcription of LCT promoter. Mutations in these regions are associated with lactose tolerance into adult life. MCM6 has been shown to interact with: Mini Chromosome Maintenance
Sources: en.wikipedia.org
Following TNF-R1 and Fas activation in mammalian cells a balance between proapoptotic (BAX, BID, BAK, or BAD) and anti-apoptotic (Bcl-Xl and Bcl-2) members of the Bcl-2 family are established. This balance is the proportion of proapoptotic homodimers that form in the outer-membrane of the mitochondrion. The proapoptotic homodimers are required to make the mitochondrial membrane permeable for the release of caspase activators such as cytochrome c and SMAC. Control of proapoptotic proteins under normal cell conditions of nonapoptotic cells is incompletely understood, but in general, Bax or Bak are activated by the activation of BH3-only proteins, part of the Bcl-2 family.
Nucleases are enzymes that cut DNA strands by catalyzing the hydrolysis of the phosphodiester bonds. Nucleases that hydrolyse nucleotides from the ends of DNA strands are called exonucleases, while endonucleases cut within strands. The most frequently used nucleases in molecular biology are the restriction endonucleases, which cut DNA at specific sequences. For instance, the EcoRV enzyme shown to the left recognizes the 6-base sequence 5′-GATATC-3′ and makes a cut at the horizontal line. In nature, these enzymes protect bacteria against phage infection by digesting the phage DNA when it enters the bacterial cell, acting as part of the restriction modification system. In technology, these sequence-specific nucleases are used in molecular cloning and DNA fingerprinting. Enzymes called DNA ligases can rejoin cut or broken DNA strands. Ligases are particularly important in lagging strand DNA replication, as they join the short segments of DNA produced at the replication fork into a complete copy of the DNA template. They are also used in DNA repair and genetic recombination.
Abelacimab (MAA868), is a fully human monoclonal antibody for the treatment of thrombosis, under development by Anthos Therapeutics. It is anti-factor XI antibody. The hope behind its development is that factor XI plays a bigger role in thrombosis (via the intrinsic/contact activation pathway) than hemostasis, so targeting it might reduce clotting risks without a corresponding increase in bleeding. It has, indeed, shown overwhelming reduction in bleeding as compared to a direct oral anticoagulant (rivaroxaban) in patients with atrial fibrillation and a moderate-to-high risk of stroke. Data is insufficient for drawing a conclusion on whether it increases or decreases stroke events as compared to rivaroxaban. Antibody Monoclonal antibody Immunology
Donald has supervised many students and postdocs, many of whom are now professors in reputed universities such as MIT, Carnegie-Mellon University, University of Washington, Seattle, University of Massachusetts Amherst, Dartmouth College, Duke University, Middlebury College and University of Toronto, and researchers at organizations such as NIAID, NIST, IBM, Sandia National Laboratories. Donald is the son of historian David Herbert Donald and historian and editor Aida DiPace Donald. MEMS Protein design NMR Protein Structure Kinodynamic planning Donald is the author of over 100 publications. A representative selection includes:
The Edman degradation is an alternative method for peptide sequencing that cleaves amino acid residues from the N-terminus of a peptide. In 1950 Edman designed a reaction with phenylthiocyanate (the idea for which was borrowed from a 1927 study by Bergmann, Kann and Miekeley ) to give phenylthiocarbamyl peptides followed by hydrolysis under relatively mild conditions to cleave N-terminal amino acid as phenylthiohydantoin. Phenylthiohydantoin is stable enough to undergo various sequencing procedures such as those which involve chromatography and mass spectrometry. This was an improvement on an earlier method proposed by Abderhalden and Brockmann in 1930 that demonstrated N-terminal amino acid conversion to a hydantoin under stronger hydrolytic conditions, where some cleavage of the residual peptide proved problematic. The primary advantage the Edman degradation has over the Bergmann degradation is the ease with which the residual peptide can re-enter the process due to retention of its structure throughout sequential cleaving. Repetition of the Bergmann degradation is presumably not as straightforward, as the remaining peptide is in amide form.
Sources: en.wikipedia.org
GCaMP consists of three key domains: an M13 domain at the N-terminus, a calmodulin (CaM) domain at the C-terminus, and a GFP domain in the center. The GFP domain is circularly permuted such that the native N- and C-termini are fused together by a six-amino-acid linking sequence, and the GFP sequence is split in the middle, creating new N- and C-termini that connect to the M13 and CaM domains. In the absence of Ca2+, the GFP chromophore is exposed to water and exists in a protonated state with minimal fluorescence intensity. Upon Ca2+ binding, the CaM domain undergoes a conformational change and tightly binds to the M13 domain alpha helix, preventing water molecules from accessing the chromophore. As a result, the chromophore rapidly deprotonates and converts into an anionic form that fluoresces brightly, similar to native GFP.
Ana María Muñoz Jauregui (born 1969, Lima) is a Peruvian pharmacist, biochemist and nutritionist. She has served as Rector and, since 2023, as the Vice Rector of research at San Ignacio de Loyola University (USIL). Muñoz has authored numerous publications and received national and international awards, such as silver, gold medals and semi-grand prize wins at the International KIWIE Award. Muñoz, was born in Lima, Peru. In 1996, she attended the National University of San Marcos (UNMSM), where she earned her pharmaceutical degree. While studying, she developed an interest in the field of nutrition. She earned a master's degree in biochemistry and nutrition, and earned a PhD in Pharmacy and Biochemistry from the National University of San Marcos (UNMSM) in 2006.
The centrifugal partition chromatograph instrument is constituted with a unique rotor which contains the column. This rotor rotates on its central axis (while HSCCC column rotates on its planetary axis and simultaneously rotates eccentrically about another solar axis). With less vibrations and noise, the CPC offers a typical rotation speed range from 500 to 2000 rpm. Contrary to hydrodynamic CCC, the rotation speed is not directly proportional to the retention volume ratio of the stationary phase. Like DCCC, CPC can be operated in either descending or ascending mode, where the direction is relative to the force generated by the rotor rather than gravity. A redesigned CPC column with larger chambers and channels has been named centrifugal partition extraction (CPE). In the CPE design, faster flow rates and increased column loading can be achieved.
Phage display libraries of 109 randomized sequences are used to screen for Affimer proteins that exhibit high-specificity binding to the target protein with binding affinities in the nM range. The ability to direct in vitro screening techniques allows the identification of specific, high affinity Affimers. In vitro screening and development also mean that the target space for Affimers is not limited by the animal immune system. Affimers are generated using recombinant systems, so their generation is more rapid and reproducible compared to the production of polyclonal antibodies. Multimeric forms Affimers have been generated and shown to yield titres in the range of 200–400 mg/L under small-scale culture using bacterial host systems. Multimeric forms of Affimers with the same target specificity provide avidity effects in target binding. Many different tags and fusion proteins, such as fluorophores, single-stranded DNA, His, and c-Myc tags can be conjugated to Affimers. Specific cysteine residues can be introduced to the protein to allow thiol chemistry to uniformly orient Affimers on a solid support eg ELISA plates. This flexible functionalisation of the Affimer molecule allows functionality across multiple applications and assay formats.
Sources: en.wikipedia.org
Method validation is the documented process of confirming that an HPLC procedure is suitable for its intended use. It evaluates accuracy, precision, specificity, linearity, range, detection limits, and robustness. Validation criteria depend on the regulatory context and the sample type.
System suitability tests are short checks performed before or during an HPLC run to verify instrument and method performance. They often include resolution, tailing factor, theoretical plates, and injection precision. Results must meet predefined limits for sample data to be accepted.
HPLC retention time alone cannot definitively identify an unknown substance. A match with a reference standard under identical conditions provides supporting evidence. Confirmation typically requires mass spectrometry, nuclear magnetic resonance, or another orthogonal technique.
It measures the amounts and identities of compounds in liquid samples by separation and detection. Depending on the detector and reference standards, results can be qualitative or quantitative. The technique is used in fields such as pharmaceutical analysis, food safety, and environmental monitoring.